Fluidwell F090 - General Purpose Indicator |
|
Features
- Universal loop-powered indicator
- Very large 26mm (1") high digits
- Displays the actual value, measuring unit and loop current
- Selectable on-screen engineering units: %, PPM, meter, m, cm, mm, ft, sqft, ML, L, NL, M3, nM3, GAL, USGAL, IGAL, bbl, CUFT, mg, g, kg, ton, lb, psi, psig, mbar, bar, barg, °C, °F, K, P, RPM, K, KPA, pH, mV, /sec, /min, /hr, /day or no unit (others on request)
- Piegraph indication: ten segments
- Number of digits: 5½
- Green/amber LED backlight with adjustable intensity
- Operational temperature -40 to 80°C (-40 to 178°F)
- Environmental protection to IP67 (NEMA 4X)
- Intrinsically Safe - ATEX, IECEx, FM and CSA approval for gas and dust applications.
Signal Input
|
Applications
The F090 takes operating energy from an input current (2-wire 4-20mA) and reduces installation cost. It accepts a wide range of process instruments, including flow (linear or square root), level, pressure and temperature transmitters. All these process parameters can be displayed in their own engineering units. The F090 is available with Intrinsic Safety for hazardous area applications.
Power Supply
|
|
OVERVIEW
Introduction
The F090 is a local indicator to display the actual process value, range and loop current. The measuring unit to be displayed is simply selected through an alphanumerical configuration menu. No adhesive labels have to be put on the outside of the enclosure: a weather proof and user friendly solution! The configuration of Span, offset and number of decimals is done through software functions, without any sensitive dip-switches or trimmers. A wide selection of options further enhance this models capabilities, including Intrinsic Safety for hazardous area applications.
Display
The display has 5½ large 26mm (1") digits to show the main process information. At the bottom line, smaller 8mm (0.31") alphanumerical digits are available to display the measuring unit and additional messages in case of an alarm; range error e.g. With the piegraph, an actual range indication is offered in a single glance. If desired, additional information can be presented after pressing the select key, like the actual loop current and the 0% and 100% process value. As the F090 has been designed for field mounted applications, a smart display update function has been incorporated to achieve a readable display even at -40°C (-40°F).
Backlight
For those applications where readability during day and night might is an issue, a bi-color backlight is available. The background color can be set to green or amber and the intensity can be adjusted from the keyboard. The display is a transflective type, which means that a high contrast reading is guaranteed in full sunlight as well as during the night. This backlight option is also available Intrinsically Safe.
Configuration
All configuration settings are accessed via a simple operator menu which can be pass-code protected. Each setting is clearly indicated with an alphanumerical description, therefore avoiding confusing abbreviations and baffling codes. Once familiar with one F-series product, you will be able to program all models in the series without a manual. All settings are safely stored in EEPROM memory in the event of sudden power failure.
Signal Input
The F090 does accept 4-20mA input signals only. The input signal type can be tuned and set by the user in the configuration menu without having to adjust any sensitive mechanical dip-switches, jumpers or trimmers.
Power Supply
The F090 is designed to be a loop-powered indicator, which means it takes its operating energy from an input current (2-wire 4-20mA).
Hazardous Area
For hazardous area applications, this model has been ATEX, IECEx, FM and CSA certified Intrinsically Safe for gas and dust applications, with an allowed operational temperature of -40°C to +70°C (-40°F to +158°F).
Enclosures
Various types of enclosures can be selected, all ATEX, IECEx, FM and CSA approved. As standard the F090 is supplied in a rugged GRP panel mount enclosure, which can be converted to an IP67 / NEMA 4X GRP field mount enclosure by the addition of a back case. Most popular is our aluminum field mount enclosure with IP67 / NEMA 4X rating. Both European or U.S. cable gland entry threads are available.
APPLICATION OVERVIEW
ENCLOSURE
Aluminum & GRP Mount Enclosure
TERMINAL CONNECTIONS
Terminal connections Power Supply PB/PC - PD - PL - PX
HAZARDOUS AREA APPLICATIONS
The F090-XI has been certified according ATEX and IECEx by KEMA and according CSA c-us and FM for use in Intrinsically Safe applications with an ambient temperature of -40°C to +70°C (-40°F to +158°F).
- The ATEX markings for gas and dust applications are:
II 1 G Ex ia IIC T4
II 1 D Ex iaD 20 IP 65/67 T 100 °C.
- The IECEx markings for gas and dust applications are: Ga Ex ia IIC T4 and Ex iaD 20 IP 65/67 T100 °C.
- The CSA c-us markings are: Class I/II/III, Division 1, Groups A, B, C, D, E, F, G, Temperature class T4 and Class I, Zone 0, AEx ia IIC T4.
- The FM markings are: Class I/II/III, Division 1, Groups A, B, C, D, E, F, G, Temperature class T4 and Class I, Zone 0, AEx ia IIC T4.
It is allowed to connect up to two I.S. power supplies to power the unit and backlight.
Configuration example IIA - IIB and IIC F090-A-PL-XI-ZB - Basic power supply 8 - 30V DC
SPECIFICATIONS
Display
Type: High intensity reflective numeric and alphanumeric LCD, UV-resistant.
Dimensions: 90 x 40mm (3.5" x 1.6").
Digits: 5½ very large 26mm (1") and eleven 8mm (0.31") digits. Various symbols and measuring units.
Piegraph: 10 segment range indication in relation to its measuring range 0 - 100%
Option ZB: Transflective LCD with bi-color LED-backlight; green / amber. Intensity and color selected through the keyboard.
Operating Temperature
Standard Unit: -40 to 80°C (-40 to 178°F).
Intrinsically Safe: -40 to 70°C (-40 to 158°F).
Power Requirements
Type PL: Input loop-powered from sensor signal 4-20mA
Type ZB: 20 - 30 VDC. Power consumption max. 1 Watt
Note For Intrinsically Safe applications, consult the safety values in the certificate
Terminal Connections
Type: Removable plug-in terminal strip. Wire max. 1.5mm2
Data Protection
Type: EEPROM backup of all settings. Data retention at least 10 years. Pass-code: Configuration settings can be pass-code protected.
Intrinsically Safe
ATEX Certification: II 1 G Ex ia IIC T4. II 1 D Ex iaD 20 IP 65 / 67 T 100 °C.
IECEx Certification: Ga Ex ia IIC T4. Ex iaD 20 IP 65 / 67 T 100 °C.
CSA c-us Certification: Intrinsically Safe for Class I/II/III, Div. 1, Groups A, B, C, D, E, F, G, Temp. class T4 and Class I, Zone 0, AEx ia IIC T4.
FM certification: Intrinsically Safe for Class I/II/III, Div. 1, Groups A, B, C, D, E, F, G, Temp. class T4 and Class I, Zone 0, AEx ia IIC T4.
Ambient: -40°C to +70°C / -40° to +158°F.
Environment
Electromagnetic compatibility: Compliant ref: EN 61326 (1997), EN 61010-1 (1993)
Signal Input
Type A: 4-20mA. Analog input signal can be scaled to any desired range within 4-20mA.
Accuracy: Resolution: 16 bit. Error < 0.01mA / ± 0.05% FS. Low level cut-off programmable
Span: 0.00001 / 199,999 with variable decimal position
Offset: -99,999 / +999,999 units
Update time: Four times per second
Voltage drop: Max. 2.6V DC @ 20mA
Relationship: Linear and square root calculation
Casing
Window: Polycarbonate window
Sealing: Silicone
Control Keys: Three industrial micro-switch keys. UV-resistant silicone keypad
Panel Mount Enclosures
Dimensions: 130 x 120 x 60mm (5.12" x 4.72" x 2.36") - W x H x D.
Panel cut-out: 115 x 98mm (4.53" x 3.86") L x H.
Type HB: Die-cast aluminum panel mount enclosure IP65 / NEMA 4.
Weight: 600 grams/1.33 lbs
Type HC: GRP panel mount enclosure IP65 / NEMA 4, UV-resistant and flame retardant.
Weight: 450 grams/1 lb
Operational
Displayed functions:
• Top line: main display value
• Bottom line: measuring unit and messages
• Piegraph: 10 segments related to input signal
• Actual sensor input value in mA.
• Displayed value at 0% of the input signal
• Displayed value at 100% of the input signal
Total
Digits: 5½ digits (height 26mm / 1").
Units: m - mL - L- nL - M3 - nM3 - mg - g - kg - ton - gal - sgal - igal - lb - bbl - cf - scf - p - rev - °C - °F - °K - % - mm - cm - mtr - inch - ft - mmwk - mmwc - cmwk - cmwc - mwk - mwc - inwc - ftwc - mbar - bar - psi - pH - mV - no unit.
Time Units: /sec - /min - /hr - /day - no unit
Decimals: 0 - 1 - 2 - 3 - 4 or 5
ORDERING INFORMATION
Standard Model Number Configuration (example): F090-A-HC-PX-XX-ZX
To aid in the construction of a Fluidwell model number a model configurator must be used. The path to this Configurator is contained in the link shown below. However, it would help to first review the options shown below.
NOTE: If you require the product to be Intrinsically Safe (ATEX, IECEx, CSA c-us or FM.), the options you choose should only include those options that carry the ATEX symbol next to them. The Configurator will not allow you to mix and match options, unless the product is to be used in a Safe Area.
Once a new model is generated using the Configurator, you will
- see the price of the fully configured model
- have an opportunity to purchase the product
- have the capability to e-mail the configuration to us, yourself, or a third person
To launch the configurator, click here
Sensor Input Signal
| A |
 |
(0)4-20mA input |
Panel Mount Enclosures - IP65 / NEMA 4
| HB |
 |
Aluminum enclosure |
| HC |
 |
GRP enclosure |
Power Supply
| PL |
 |
Input loop-powered from sensor signal 4-20mA (type A) |
Hazardous Area
| XI |
 |
Intrinsically Safe |
| XX |
|
Safe area only |
Other Options
| ZB |
 |
Backlight |
| ZX |
 |
No options |
The bold marked text contains the standard configuration.
|
DOCUMENTATION
|