Loop-Powered General Purpose Indicator F090
Features General Information Application Overview Dimensions

Enclosures Terminals Specifications Ordering Information

  Configurator Documentation  




Features

  • Universal loop-powered indicators
  • Very large 26mm (1") high digits
  • Display the actual value, measuring units and loop current
  • Selectable on-screen engineering units: %, PPM, meter, m, cm, mm, ft, sqft, ML, L, NL, M3, nM3, GAL, USGAL, IGAL, bbl, CUFT, mg, g, kg, ton, lb, psi, psig, mbar, bar, barg, °C, °F, K, P, RPM, KPA, /sec, /min, /hr, /day or no unit (others on request)
  • Piegraph indicator: ten segments
  • Number of digits: 5½
  • Green/amber LED backlight with adjustable intensity
  • Operational temp: -40 to 80°C (-40 to 178°F)
  • Environmental protection to IP67 (NEMA 4X)
  • Intrinsically Safe
    II 1 GD EEx ia IIC T4 T100°C




Loop-Powered General Purpose Indicator F090

Signal Input

  • 4-20mA

Power Supply

  • Loop-Powered

Applications

  • The F090 takes operating energy from an input current (2 wire 4-20mA) and reduces installation cost. It accepts a wide range of process instruments, including flow (linear or square root), level, pressure and temperature transmitters. All these process parameters can be displayed in their own engineering units. The F090 is, like other F-Series, optional available with Intrinsic Safety for hazardous area applications.


General Information

Introduction

The F090 is a local indicator to display the actual process value, range and loop current. The measuring units to be displayed is simply selected through an alphanumeric configuration menu. No adhesive labels have to be put on outside of the enclosure: a weather proof and user friendly solution! The configuration of Span, offset, and number of decimals done through software functions, without any sensitive dip-switches or trimmers. A wide selection of options further enhance this models capabilities, including Intrinsic Safety for hazardous area applications.

Display

The display has 5½ large 26mm (1") digits to show the main process information. At the bottom line line, smaller 8mm (0.31") alpha-numericial digits are available to display the measuring units and additional messages in case of an alarm; range error e.g. With the piegraph, an actual range indication is offered in a single glance. If desired, additional information can be presented after pressing the select key, like the actual loop current and 0% and 100% process value. As the F090 has been designed for field mounted applications, a smart display update function has been incorporated to achieve a readable display even at -40°C (-40°F).

Backlight

For those applications where readability during day and night is an issue, a bi-colored backlight is available. The background color can be set to green or amber and the intensity can be adjusted from the keyboard. The display is a transflective type, which means that high contrast reading is guaranteed in full sunlight as well as during the night. This backlight option is available Intrinsically Safe.

Configuration

All configuration settings are accessed via a simple operator menu which can be pass-code protected. Each setting is clearly indicated with an alphanumeric description, therefore avoiding confusing abbreviations and baffling codes. Once familiar with one F-Series products, you will be able to program all models in the series without a manual. All settings are safely stored in EEPROM memory in the event of sudden power failure.

Signal Input

The F090 does except 4-20mA signals only. The input signal type can be turned and set by the user in the configuration menu without having to adjust any sensitive mechanical dipswitches, jumpers or trimmers.

Power Supply

The F090 is designed to be loop powered indicator, which means it takes its operating energy for an input current (2-wire 4 - 20mA)

Hazardous Area

For hazardous area applications, this model has been ATEX certified Intrinsically Safe II 1 GD EEx ia IIC T4 T100C with an allowed operational temperature of -40 to 70°C (-40 to 158°F). IEC, CSA and FM certification is expected to be available in Q3 2007.

Enclosures

Various types of enclosures can be selected, all ATEX approved. As standard the F090 is supplied in a rugged GRP panel mount enclosure, which can be converted to an IP67/NEMA4X GRP field mount enclosure by the addition of a back case. Most popular is our aluminum field mount with IP/67 NEMA 4X rating. Both European or US cable gland threads are available


Loop-Powered General Purpose Indicator  Application Overview

Dimensions

Aluminum & GRP Mount Enclosure

Loop-Powered General Purpose Indicator Enclosures

Aluminum Field/Wall Mount Enclosures

Loop-powered General indicator Fluidwell F090 Enclosures

Loop-Powered General Purpose Indicator HA Enclosure
HA
Loop-Powered General Purpose Indicator HM Enclosure
HM
Loop-Powered General Purpose Indicator HN Enclosure
HN
Loop-Powered General Purpose Indicator HO Enclosure
HO
Loop-Powered General Purpose Indicator HP Enclosure
HP
Loop-Powered General Purpose Indicator HT Enclosure
HT
Loop-Powered General Purpose Indicator HU Enclosure
HU
Loop-Powered General Purpose Indicator HZ Enclosure
HZ
GRP Field/Wall Mount Enclosures
Loop-Powered General Purpose Indicator HD Enclosure
HD
Loop-Powered General Purpose Indicator HE Enclosure
HE
Loop-Powered General Purpose Indicator HF Enclosure
HF
Loop-Powered General Purpose Indicator HG Enclosure
HG
Loop-Powered General Purpose Indicator HH Enclosure
HH


Terminal Connections

Power Supply Type PL

Loop-powered General indicator Fluidwell F090 Terminal Connections
Display

90x40mm (3.5" x 1.6")

Loop-powered General indicator Fluidwell F090 Display
Hazardous Area Applications

The F090-XI has been ATEX approved by KEMA for use in Intrinsically Safe applications. It is approved according to II 1GD EEx ia IIC T4 T100°C for gas and dust applications with an operational range of -40°C to +70°C (-40°F to +158°F). IEC, CSA and FM approvals are expected to become available in Q3 2007. It is allowed to connect up to two I.S. power supplies to power the unit and backlight. Please contact your supplier for further details.


Certificate of Conformity KEMA 05ATEX1168 X

Loop-powered General indicator Fluidwell F090 Certificate of Conformity
Configuration example IIA - IIB and IIC
F090-A-PL-XI-XB Input Loop-Powered

Loop-powered General indicator Fluidwell F090 Configuration

Technical Specifications
General
Display
Type High intensity reflective numeric and alphanumeric LCD, UV-resistant
Dimensions 90 X 40 mm (3.5" X 1.6")
Digits 5½ very large 26mm (1") and eleven 8mm (0.31") digits. Various symbols and measuring units
Piegraph 10 segment range indication in relation to its measuring range 0-100%
Option ZB Transflective LCD with bi-color LED backlight; green / amber. Intensity and color selected through the keyboard
 
Operating Temperature
Standard Unit -40 to 80°C (-40 to 178°F)
Intrinsically Safe -40 to 70°C (-40 to 158°F)
 
Power Requirements
Type PL Input loop-powered from sensor 4-20mA
Type ZB For backlight 20-30 VDC
Power consumption max 1 watt
Note For Intrinsically Safe applications, consult the safety values in the certificate
 
Terminal Connection
Type Removable plug-in terminal strip. Wire max. 1.5mm2
 
Data Protection
Type EEPROM backup of all settings. Data retention at least 10 years
Pass-code Configuration settings can be pass-code protected
 
Hazardous Area
Intrinsically Safe
ATEX Certification II 1GD EEx ia IIC T4 T100°C
CSA C-US/IECEx Certification IEC, CSA, and FM Approvals are expected to become available in Q3 2007
 
Environment
Electromagnetic Compatibility Compliant ref: EN 61326 (1997), EN 61010-1 (1993)
 
Signal Input
Sensor
Type A 4-20 mA. Analog input signal can be scaled to any desired range within 4-20mA
Accuracy Resolution: 16 bit. Error < 0.01 mA / ± 0.05 % FS.
Low level cut-off programmable
Span 0.001 to 99,999 with variable decimal position
Offset -99,999 to +99,999 units
Update time Four times per second
Voltage drop Max. 2.6 VDC at 20 mA
Relationship Linear and square root calculations
 
Casing
General
Window: Polycarbonate window
Sealing: Silicone
Control Keys: Three industrial micro-switch keys, UV resistant silicone keypad
Option ZB: Silicone free ABS enclosure with EPDM and PE sealings. UV-resistant polyester keypad
Note: This option comes with HD only
 
Aluminum Wall/ Field Enclosures
General Die-cast aluminum wall/field mount enclosure IP67/NEMA4X with 2 component UV-resistant coating
Dimensions 130 X 120 X 75mm (5.12" X 4.72" X 2.95") - W X H X D
Weight 37 1/2 ounces
Type HA Cable Entry: 2 x PG9 and 1 x M20
Type HM Cable Entry: 2 x M16 and 1 x M20
Type HN Cable Entry: 1 x M20
Type HO Cable Entry: 2 x M20
Type HP Cable Entry: 6 x M12
Type HT Cable Entry: 1 x ½" NPT
Type HU Cable Entry: 3 x ½" NPT
Type HZ Cable Entry: No Holes
 
GRP Wall/Field Mount Enclosures
General GRP wall/field mount enclosure IP67/NEMA 4X, UV-resistant and flame retardant
Dimensions 130 X 120 X 75mm (5.12" X 4.72" X 2.95") - W X H X D
Weight 20 ounces
Type HD Cable Entry: No Holes
Type HE Cable Entry: 2 x 0 16mm and 1 x 0 20mm
Type HF Cable Entry: 1 x 0 22mm (0.866")
Type HG Cable Entry: 2 x 0 20mm
Type HH Cable Entry: 6 x 0 12mm
 
Panel Mount
Dimensions 130 X 120 X 75mm (5.12" X 4.72" X 2.95") - W X H X D
Panel cut-out 115 x 98mm (4.53" x 3.86") L x H
Type HB Die-cast aluminum panel mount enclosure IP65/NEMA 4
Weight 570 grams
Type HC GRP panel mount enclosure IP65/NEMA 4
UV-resistant and flame retardant
Weight 15 ounces
 
Operational
Displayed Functions  Top line: main display value
 Bottom Line: measuring units and messages
 Piegraph: 10 segment display related to input signal
 Actual sensor input value in mA
 Display value at 0% of the input signal
 Display value at 100% of the input signal
 
Total
Digits 5½ digits (height 26mm / 1")
Units m mL - L - nL - M3 - nM3 - g - kg - ton - gal - sgal - igal - lb - bbl - cf - scf - p - rev - °C - °F - °K - % - mm - cm - mtr - inch - ft - mmwk - mmwc - cmwk - cmwc - mwk - mwc - inwc - ftwc - mbar - bar - psi - no unit
Time Units /sec, /min, /hr, /day - no unit
Decimals 0 - 1- 2 or 3
 
Piegraph
Digits 10 segments
Relation to the min. and max. input signal (0 - 100%)
   

Ordering Information

Standard Configuration
Ordering Information: F090 -A -H_ -PL -X_ -Z_
Sensor Input        
A   Loop-powered General indicator Fluidwell F090 Atex Symbol  4-20 mA Input
Panel Mount Enclosures - IP65/NEMA 4
HB  Loop-powered General indicator Fluidwell F090 Atex Symbol  Aluminum Enclosure
HC  Loop-powered General indicator Fluidwell F090 Atex Symbol  GRP Enclosure
GRP Wall/Field mount Enclosures - IP67/NEMA 4x
HD  Loop-powered General indicator Fluidwell F090 Atex Symbol  Cable Entry: No Holes
HE  Loop-powered General indicator Fluidwell F090 Atex Symbol  Cable Entry: 2 x 16mm + 1 x 20mm
HF  Loop-powered General indicator Fluidwell F090 Atex Symbol  Cable Entry: 1 x 22mm (0.866")
HG  Loop-powered General indicator Fluidwell F090 Atex Symbol  Cable Entry: 2 x 20mm
HH  Loop-powered General indicator Fluidwell F090 Atex Symbol  Cable Entry: 6 x 12mm
Aluminum Field/ Wall Mount Enclosures IP67/NEMA 4X
HA  Loop-powered General indicator Fluidwell F090 Atex Symbol  Cable Entry: 2 x PG9 + 1 x M20
HM  Loop-powered General indicator Fluidwell F090 Atex Symbol  Cable Entry: 2 x M16 + 1 x M20
HN  Loop-powered General indicator Fluidwell F090 Atex Symbol  Cable Entry: 1 x M20
HO  Loop-powered General indicator Fluidwell F090 Atex Symbol  Cable Entry: 2 x M20
HP  Loop-powered General indicator Fluidwell F090 Atex Symbol  Cable Entry: 6 x M12
HT  Loop-powered General indicator Fluidwell F090 Atex Symbol  Cable Entry: i x ½" NPT
HU  Loop-powered General indicator Fluidwell F090 Atex Symbol  Cable Entry: 3 x ½" NPT
HZ  Loop-powered General indicator Fluidwell F090 Atex Symbol  Cable Entry: No Holes
Power Supply
PL  Loop-powered General indicator Fluidwell F090 Atex Symbol  Input Loop-Powered from a sensor signal 4-20 mA (type A)
Hazardous Area
XI  Loop-powered General indicator Fluidwell F090 Atex Symbol  Intrinsically Safe
XX    Safe Area only
Other Options
ZB  Loop-powered General indicator Fluidwell F090 Atex Symbol  Backlight
ZS  Loop-powered General indicator Fluidwell F090 Atex Symbol  Silicone free ABS enclosures with EPDM and PE sealings (type HD only)
ZX  Loop-powered General indicator Fluidwell F090 Atex Symbol  No Options

Fluidwell Configurator

See the Fluidwell F090 General Purpose Loop-powered Indicator using the Configurator for pricing information, click here
 
Documentation

The documentation can be downloaded from these links:
F090: Data Sheet
Fluidwell Product Information
Low-Cost Explosion Proof Enclosure
Extended Warranty Info
Back to Top

See How You can Use Precision Digital Products for
 What's New | News | Products | Pricelist | Services | Support | Jobs | Contact Us | Site Map || Admin
Precision Digital: Distributors and Manufacturers of Digital Pnale Meters and Indicators